Hello, I have a problem with OCT setup I am trying to make. The setup is based on Michelson interferometer with mirrors in the sample and the reference arms. Light source is 639 nm high-power (up to 150 mW) diode laser running in CW mode with coherence length ~150 microns. After the laser I have non-polarizing cubic beamsplitter (50%/50%). The mirror in reference arm is modulated by PZT (2.3 kHz, ~375 nm amplitude). After the reference and sample beams combined I have high-speed photodetector connected to the oscilloscope with 50 Ohms termination load. The photodetector is DET36A/M provided by Thorlabs (rise time ~14 ns, 350-1100 nm wavelength range, 13 mm2 active area). I am supposed to get only one reference mirror position when I have constructive interference. In the case of constructive interference on the oscilloscope I see periodic 6-8 Intensity oscillations (mirror goes forward and backward) with Doppler frequency. The problem is in reality I have repetitive constructive interferences ~ every 8.8 mm (reference mirror position is 0 mm, 8.8 mm, 17.6 mm, etc.). And the oscillation amplitude on the oscilloscope is almost the same. Can you please give me any suggestions/comments what can be wrong in the setup? Thanks!