peacefulspring Posted April 17, 2009 Posted April 17, 2009 Hi, Can anyone kindly tell me how to quantify DNA other than using spectrophotometer/ gel electrophoresis. And how could we measure DNA using only 1microlitre? Thanks.
CharonY Posted April 17, 2009 Posted April 17, 2009 Well you could do it with a nanodrop (only needs 1µl), but is also a spectralphotometric method. Alternatives are a fluorescence based assay with subsequent measurement with a fluorometer (e.g. picogreen). You could also make isotope dilution measurements with a MS, but that is kind of overkill. 1
CharonY Posted April 21, 2009 Posted April 21, 2009 Well, yeah, but you would need identical DNA as a reference standard to be able to measure abundance.
peacefulspring Posted April 21, 2009 Author Posted April 21, 2009 But if I use Ethidium Bromide Fluorescence for the amplifiable DNA from maize (For the detection of GM maize experiement), is it alright too?
CharonY Posted April 21, 2009 Posted April 21, 2009 It is possible but : - it is overall rather inaccurate - you need a reference standard in the same run with about the same length of the fragment(s) you want to quantify (as the incorporation of EtBr is length dependent)
peacefulspring Posted April 22, 2009 Author Posted April 22, 2009 Thus, if I choose to use Fluorometric Quantitation for my maize's DNA, would it be better? Merged post follows: Consecutive posts mergedbesides that, usually for negative control in detection for maize dna for instance, is the content of the negative control containing 1) only distilled water 2) everything (e.g. primer,polymerase & stuffs) except for DNA
CharonY Posted April 22, 2009 Posted April 22, 2009 Technically yes, as methods as e.g. picrogreen have a higher level of linearity (around three orders of magnitude). You still need to make a standard curve, though (is true for every quantification technique, actually). The length dependency is also a bit lower than for EtBr. And 2 is the way to go. For controls you should have precisely the same parameters except the one you are testing.
peacefulspring Posted May 12, 2009 Author Posted May 12, 2009 besides that, what would be the optimum voltage for 2% agarose gel and the duration for it? In addition, is the pore size of 2% agarose bigger than 1% agarose? Merged post follows: Consecutive posts mergedin addition to the previous post, the amplified products ranges form 100-600bp in size.. can anymore rly asap cos it's urgent!thanks..
lucaspa Posted May 12, 2009 Posted May 12, 2009 besides that, what would be the optimum voltage for 2% agarose gel and the duration for it? In addition, is the pore size of 2% agarose bigger than 1% agarose? The pore size of a 2% agarose gel would be smaller than a 1% agarose gel. More agarose, denser the agarose molecules and the smaller the pores between them. No matter the voltage, you can track the migration of DNA thru the gel with the dyes bromphenol blue and xylene cyanol. http://www.vivo.colostate.edu/hbooks/genetics/biotech/gels/agardna.html However, "As the voltage applied to a gel is increased, larger fragments migrate proportionally faster that small fragments. For that reason, the best resolution of fragments larger than about 2 kb is attained by applying no more than 5 volts per cm to the gel (the cm value is the distance between the two electrodes, not the length of the gel). " I would suggest you do 5 volts per cm. 1
CharonY Posted May 12, 2009 Posted May 12, 2009 (edited) If you want to have good resolution down to 100 bp you should definitely use 1.8-2% gels. Regarding voltages, it also depends on buffer capacity. As a rule of thumb, lower voltages generally yield better resolution. However the trick is also to stop the gel at the right time, depending on where you expect the majority of bands. Edit: many minigels have a size of 8-7 cm. But running at 5V/cm (or 40-35 V in total) is not really necessary. Usually anything from 50-80 V generally works pretty well. Edited May 12, 2009 by CharonY
peacefulspring Posted May 13, 2009 Author Posted May 13, 2009 I've another question regarding DNA extraction from maize. So can anyone kindly teach me how to make fresh kernels to powder form? Is it by using liquid nitrogen so that the corn powder that we obtained would be to be extracted? And after which, can we simply use food mixer/ pestle/ mortar to 'grind' it to the powder form that we want? Thanks.
Recommended Posts
Create an account or sign in to comment
You need to be a member in order to leave a comment
Create an account
Sign up for a new account in our community. It's easy!
Register a new accountSign in
Already have an account? Sign in here.
Sign In Now