So in my bio class we had our first lab as a microscopy lab. Since I am a transfer student, they want me to take freshman bio even though I got a 5 on AP exam which exempt me from freshman bio at my other institution. I state that because I went in all cocky to this lab since I have used microscopes many times before this in my other classes such as Invertebrate Zoology. In that we had a microscopy lab, but we never dealt with a stage micrometer for some odd reason. When this was foreign to me, I felt stressed and embarrassed for myself and the instructor (he was going nuts trying to explain things to people about it) . So, since I am new to it, I want to know if these calculations are correct...

 

At 4x objective, if 25 stage units fit into 10 ocular units, then is that equal to .25 mm and 250 um? I just did 25x.01mm and then got my answers. I feel like that's wrong.

 

From an earlier determination that 2.5 stage units fit into 10 ocular units at 40x objective, I calculated that 80 ocular units is the equivalent to 320 um. I did this by doing 80/2.5, multiplying that by .01 and then converting mm to um. Correct?

 

I hope I'm right, I want to stop feeling like an idiot over such a simple thing

 

Thanks in advance.

Link to comment
Share on other sites

So in my bio class we had our first lab as a microscopy lab. Since I am a transfer student, they want me to take freshman bio even though I got a 5 on AP exam which exempt me from freshman bio at my other institution. I state that because I went in all cocky to this lab since I have used microscopes many times before this in my other classes such as Invertebrate Zoology. In that we had a microscopy lab, but we never dealt with a stage micrometer for some odd reason. When this was foreign to me, I felt stressed and embarrassed for myself and the instructor (he was going nuts trying to explain things to people about it) . So, since I am new to it, I want to know if these calculations are correct...

 

At 4x objective, if 25 stage units fit into 10 ocular units, then is that equal to .25 mm and 250 um? I just did 25x.01mm and then got my answers. I feel like that's wrong.

 

I believe this is correct. 25 stages/10 ocular units * 0.1 mm/stage = .25 mm or 250 um per ocular unit at 4x.

 

From an earlier determination that 2.5 stage units fit into 10 ocular units at 40x objective, I calculated that 80 ocular units is the equivalent to 320 um. I did this by doing 80/2.5, multiplying that by .01 and then converting mm to um. Correct?

 

25 um / OU at 40x * 80 OU = 2.0 mm You just need to watch the units to help you avoid dividing when you mean to multiply.

 

I hope I'm right, I want to stop feeling like an idiot over such a simple thing

 

Thanks in advance.

 

It seems like you understand how it works.

Link to comment
Share on other sites

Yes, but try not to do it from memory, instead reason through the problem based on the information provided.

 

In the second case you have 2.5 stage units / 10 OU and each stage unit is 0.1 mm so --->

 

(2.5 SU/10 OU) * 0.1 mm/SU = 0.025 mm/OU You know to multiply because the stage units must cancel this gives you the distance per ocular unit and since you have 80 OU you again multiply so the OU's cancel to get total length of 2.0 mm.

Link to comment
Share on other sites

Yes, but try not to do it from memory, instead reason through the problem based on the information provided.

 

In the second case you have 2.5 stage units / 10 OU and each stage unit is 0.1 mm so --->

 

(2.5 SU/10 OU) * 0.1 mm/SU = 0.025 mm/OU You know to multiply because the stage units must cancel this gives you the distance per ocular unit and since you have 80 OU you again multiply so the OU's cancel to get total length of 2.0 mm.

 

If the total length is 2.0 mm, that means that the Paramecium is 2000 micrometers. That seems pretty false...what am I not grasping here?

Link to comment
Share on other sites

Let's look at it another way to confirm. If the stage glass is in 0.1 mm graduations, and ten ocular units spans 2.5 stages then 10 OU is .25 mm so eight times .25 mm is still 2.0 mm. so 80 OU at 40x is 2 mm unless the stage glass is not 0.1 mm.

Link to comment
Share on other sites

Ah, you're totally right man. I found a formula:

 

one space on the OM= X spaces on SM x .01mm/ 10

 

Thanks so much for your help.

 

So are these all right? I found an easy formula for the last part too

 

Number of SM units in 10 OU @4X: 25 One OM unit= .025mm,25um

Number of SM units in 10 OU @10X:10 One OM unit=.01mm, 10um

Number of SM units in 10 OU @40X: 2.5 One OM unit=.0025mm, 2.5um

 

By doing (length of one division of the ocular micrometer)(number divisions counted as length Paramecium caudatum).

So (80)(2.5)

 

80 OU divisions @ 40X= 200um

70 OU divisions @ 40X= 175um

Edited by Marconis
Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.